Omron Forums Support Posted March 14, 2022 Share Posted March 14, 2022 This is a short description of the functionality of the Photometry settings upon startup. The Photometry settings are gain, exposure and black level offset (ADC). These are typically set with the Normalization program. The Normalization program picks photometry settings that will allow calibration to be successful using our standard calibration target. Upon startup, the photometry settings are read from the current settings for the CONTRASTCAL if the unit has no verification license installed or the XLIGHTCONTRASTCAL if there is a verification license installed. The parameters of CONTRASTCAL are black offset (ADC), exposure, gain The parameters of XLIGHTCONTRASTCAL are light, black offset (ADC), exposure, gain To get the HE1500 to start with a given photometry settings, set that photometry setting using either CONTRASTCAL or XLIGHTCONTRASTCAL Typical usage would be: XLIGHTCONTRASTCAL 90 0 2500 200 Quote Link to comment Share on other sites More sharing options...