Omron Forums Support Posted March 14, 2022 Posted March 14, 2022 White Balancing TMGC03C and TMG03C cameras 1) Shut down FrontRunner. 2) Shut down AvpServer 3) Delete dmconfig.dm file from “c:\vscape\dm” folder 4) Start Visionscape GigE Camera Control software 5) Make yourself Guru 6) Select Camera 7) Change Pixel Mode to “BGR8Packed” 8) Start Live Video 9) Go to g) Analog Controls 10) Set camera up to look at white paper with your light 11) Execute White Balance - This sets the R,G and B analog gains on the camera. 12) Change Pixel Mode back to BayerRGB 13) Go to j) UserSets 14) Select UserSet1 instead of Default 15) Make UserSetDefaultSelector UserSet1 as well so it powers up with new settings 16) Execute “UserSetSave” to save settings to camera. 17) Camera should now be set to power up in BayerRGB mode with R,G and B Analog Gains set for White Balance Quote